Special requirements of fiber optic sensing systems (FBG) for patch cords?

Why must APC connectors be used to suppress background noise?

In Fiber Bragg Grating (FBG) sensing systems, the grinding process of jumper connector end-faces critically impacts the system’s Signal-to-Noise Ratio (SNR) and measurement stability. The general requirement for systems to employ APC (Angled Physical Contact) connectors instead of PC/UPC (Physical Contact/Ultra Physical Contact) connectors stems from the core physical principle of utilizing the angled facet to eliminate axial Fresnel reflection, thereby suppressing background noise and parasitic interference.


I. Physical Mechanism Analysis: Reflection Differences Between PC/UPC and APC

1. Fresnel Reflection

When light propagates across the interface of media with different refractive indices (e.g., fiber core n_1 \approx 1.468 and air in the end-face gap n_2 \approx 1.0), Fresnel reflection inevitably occurs. The power reflectivity R at normal incidence is determined by the following formula:

R = \left( \frac{n_1 - n_2}{n_1 + n_2} \right)^2

Under ideal air gap conditions, the Fresnel reflectivity at this interface is approximately 4\% (around -14\text{ dB}). Even with UPC connectors, which feature high-precision polishing for physical end-face contact, microscopic irregularities on the order of micrometers or slight air gaps are unavoidable, resulting in a return loss typically only between -50\text{ dB} and -55\text{ dB}.

2. Conduction Path of Reflected Light Modes

  • PC / UPC End-Face (0^\circ Flat Polish):
    The normal to the end-face is parallel to the fiber core axis. When Fresnel reflection occurs, the reflected light travels back along the original optical axis. Since its angle of incidence falls within the fiber’s Numerical Aperture (NA), the vast majority of the reflected light is re-coupled directly into the core and returns to the demodulation system as a guided mode.
  • APC End-Face (8^\circ Angled Polish):
    The end-face is manufactured with a standard 8^\circ angle. According to the law of reflection, when light reflects at the fiber/air or contact interface, the reflected ray deviates from the fiber core axis by 16^\circ. This angle significantly exceeds the critical angle for total internal reflection in standard single-mode fibers (where the angle of incidence is less than the critical angle between the core and cladding interface). Consequently, the reflected light cannot be confined within the core and instead leaks into the cladding and is absorbed by the outer coating.
    Therefore, APC connectors typically achieve a return loss better than -60\text{ dB} or even -65\text{ dB}, suppressing the return reflection power by several orders of magnitude.

II. Stringent Requirements for APC Connectors in FBG Sensing Systems

While communication systems primarily focus on insertion loss, FBG sensing systems employ demodulators that function as highly sensitive, wide dynamic range reflection spectrum analyzers. Axial reflections from flat end-faces can cause severe signal degradation:

  1. Masking Weak Grating Signals and Elevating System Noise Floor
    FBG sensing technology relies on detecting minute shifts in the center wavelength \lambda_B of the reflected fiber grating. In multiplexed networks or systems with Weak Fiber Bragg Gratings (WFBGs), the reflected power from a single grating can be less than 1\% of the incident light ( < -20\text{ dB} or even lower). If a PC end-face on the link introduces a broadband reflection of -40\text{ dB} or higher, this reflected light superimposes on the detector, significantly raising the baseline noise floor and severely compressing the spectral SNR, leading to weak signals being submerged by noise.

  2. Preventing Parasitic Fabry-Pérot (F-P) Interference
    If two or more flat-end connectors (such as PC connectors at both ends of a jumper) exist in the link, they can form a parasitic, low-finesse F-P resonant cavity. When a broadband light source or a swept-frequency laser is incident, noticeable periodic interference ripples will appear in the spectrum. These ripples, superimposed on the FBG’s reflection peak, can cause demodulation algorithms (like peak-finding or Gaussian fitting algorithms) to compute incorrectly, leading to pseudo-drifts and random jitter in the measurement data.

  3. Protecting Demodulator Internal Light Source and Detector Modules
    Narrow-linewidth active devices such as Tunable Laser Sources (TLS) are extremely sensitive to back reflections. Strong end-face reflections entering the laser cavity can lead to linewidth broadening, degradation of Relative Intensity Noise (RIN), or even multi-mode hopping, causing misalignment of the demodulation light source’s wavelength output.


III. Official Technical Specifications

To ensure the sensing link exhibits extremely low back reflection and high stability, Beijing Dacheng Yongsheng Technology Co., Ltd.'s OFSCN® series of fiber patch cords are manufactured by default with FC/APC connectors (other types like SC/APC and LC/APC can also be customized).