How can customers determine if the entire system installation is qualified?
After the installation of fiber optic sensing systems (such as Fiber Bragg Grating (FBG) sensing systems or Distributed Fiber Optic Sensing (DOFS) systems) in the field, engineering acceptance typically involves a comprehensive evaluation based on four core indicators: optical link integrity, spectrum and signal quality, initial baseline consistency of physical quantities, and system operational stability:
1. Optical Link Integrity and Attenuation Test (Optical Link Loss)
- Total Link Attenuation: Use an Optical Time Domain Reflectometer (OTDR) or an optical power meter to measure the loss of the fiber transmission path and connectors. The insertion loss of individual active connectors (e.g., FC/APC) should comply with engineering specifications (typically requiring \le 0.5\text{ dB} loss), and the overall link attenuation must reserve a demodulation safety margin of 3\text{ dB} \sim 6\text{ dB} .
- End-face and Reflectance Inspection: Connector end-faces should be clean, free of contamination and scratches. For APC (Angled Physical Contact) connectors, the end-face return loss must meet specifications (typically \ge 60\text{ dB} ) to prevent strong reflected light from saturating the demodulation optical path or generating spurious peaks.
2. Spectral and Signal Quality Verification (Spectral & Signal Integrity)
- Signal-to-Noise Ratio (SNR) and Peak Symmetry: Examine the reflected or transmitted spectrum on the demodulator host. A qualified FBG sensing signal should exhibit a clear, steep, and symmetrical spectral peak. The SNR typically needs to be \text{SNR} \ge 15\text{ dB} . The spectrum should not show splitting (birefringent double peaks due to polarization), top flattening distortion, or severe attenuation caused by small bending radii ( R < 15\text{ mm} ) or illegal squeezing.
- Channel Wavelength Non-Overlap: For wavelength division multiplexing (WDM) grating string systems, the center wavelengths of each sensing measurement point must be arranged at designed intervals to ensure no wavelength overlap within the full range of physical quantity variations.
3. Initial Physical Quantity and Factory Calibration Comparison (Baseline & Calibration Consistency)
- Initial Center Wavelength Verification: Under known ambient temperature and without abnormal external force constraints, record the initial center wavelength \lambda_0 after sensor installation.
- Calibration Formula Deviation Verification: Compare the initial center wavelength \lambda_0 measured on-site with the reference wavelength from the factory calibration certificate. When installed without prestress, the deviation should be within the factory tolerance range. If prestressed or pre-tensioned during installation, the wavelength shift must match the calculated value for the designed pre-tension force. Abnormal wavelength deviations usually indicate excessive stretching, twisting, or debonding of the sensor.
4. Mechanical Fixation and Protection Effectiveness (Mechanical Fixation & Protection)
- Fixation and Strain Transfer Efficiency: For strain sensors, such as the Beijing Dacheng Yongsheng Technology Co., Ltd. OFSCN® Alloy Tube Packaged Fiber Bragg Grating strain sensor, the fixture welding, bolt tightening, or adhesive layer must be firm and free from looseness, ensuring that strain is transferred to the internal grating without loss.
- Physical Protection Standards: Inspect the sensor’s packaging tube and armored transmission cable (e.g., the seamless steel pipe structure of the Beijing Dacheng Yongsheng Technology Co., Ltd. OFSCN® 300°C Fiber Bragg Grating Temperature Sensor) for integrity. Ensure that the waterproof, dustproof, temperature-resistant, and insulation protective layers meet the requirements of the harsh on-site operating conditions.
5. System Static Noise and Continuous Operation Stability (System Stability & Noise Floor)
- Static Noise Floor Test: With the structure under test in a static state and the ambient temperature and humidity relatively constant, enable continuous system sampling (typically for 1 to 24 hours) and calculate the standard deviation \sigma of the wavelength or physical quantity. The system’s static wavelength noise should comply with the technical standards of the demodulator (e.g., \sigma \le 1\text{ pm} ).
- Repeatability and Step Response: Apply a known physical quantity change to the measurement point (e.g., a known temperature difference change or loading of a standard weight). The system’s output value should respond quickly and show no residual zero drift after recovery. Both the return error and repeatability error must be controlled within the designed allowable limits.